Hello Oddgeir如果您有E5061B选项3L5,它具有独特的过载保护功能。
当E5061B 3L5检测到过载时,它会改变仪器设置并保护内部威廉希尔官方网站
。
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm您需要使用SCPI命令清除过载保护状态“:SERVICE:POWER:OVERload:PROTect:CLEar”http:
//ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm但我担心,如果你有这种情况,Labview应用驱动程序可能会阻止该工具。
如果您无法向E5061B发送任何SCPI命令,请联系NI支持团队以了解如何重置被阻止的通信。
以上来自于谷歌翻译
以下为原文
Hello Oddgeir
If you have E5061B option 3L5, this has unique overload protection functions.
When E5061B 3L5 detects overload, it changes instrument setting and protect internal circuit.
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm
You need to clear overload protection status with a SCPI command ":SERVice:POWer:OVERload:PROTect:CLEar"
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm
But I am afraid that the instrument may be blocked by Labview appliction of driver in the case you have. If you can't send any SCPI command to E5061B, please contact NI support team to know how to reset the blocked communication.
Hello Oddgeir如果您有E5061B选项3L5,它具有独特的过载保护功能。
当E5061B 3L5检测到过载时,它会改变仪器设置并保护内部威廉希尔官方网站
。
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm您需要使用SCPI命令清除过载保护状态“:SERVICE:POWER:OVERload:PROTect:CLEar”http:
//ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm但我担心,如果你有这种情况,Labview应用驱动程序可能会阻止该工具。
如果您无法向E5061B发送任何SCPI命令,请联系NI支持团队以了解如何重置被阻止的通信。
以上来自于谷歌翻译
以下为原文
Hello Oddgeir
If you have E5061B option 3L5, this has unique overload protection functions.
When E5061B 3L5 detects overload, it changes instrument setting and protect internal circuit.
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm
You need to clear overload protection status with a SCPI command ":SERVice:POWer:OVERload:PROTect:CLEar"
http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm
But I am afraid that the instrument may be blocked by Labview appliction of driver in the case you have. If you can't send any SCPI command to E5061B, please contact NI support team to know how to reset the blocked communication.