这变得更加有趣!
锁定角度电位器调整肯定存在于软件的ESG-B端(例如,如果我告诉它我有E4432B),但ESG-AP / DP端不存在(当我告诉它时)
我实际拥有的内容:E4436B)。在我的服务支持软件(版本B.07.00,ESG_B版本B7.0,ESG_APDP版本B6.3)中选择了“调整”单选按钮,并选择了E4436B仪器,顺序为
测试是:ANALOG BUS ADCPRETUNE CAL.VCO BIAS POT ADJ.INTERNAL SOURCE CAL.KV VS FREQUENCY CAL.TIME BASE DAC CAL.FM SCALE DAC OFFSET CAL ....等这是软件的问题,或者是
ESG-AP / DP仪器中不需要锁定角度调整?
我仔细查看了校准指南,找不到任何表明不应该或不能在ESG-AP / DP仪器上进行此调整的内容。
有解决方法吗?
有没有办法告诉软件在我的E4436B上运行这个调整?
(例如,通过告诉软件该仪器是E4432B,或以某种方式将Lock Angle Pot测试安装到软件的AP / DP“侧”来说谎软件。)问候,Stan
以上来自于谷歌翻译
以下为原文
This is becoming even more intriguing! The Lock Angle Potentiometer adjustment is definitely present in the ESG-B side of the software (if, for example, I tell it I have an E4432B), but it is not present in the ESG-AP/DP side (when I tell it what I actually have: an E4436B).
In my Service Support Software (Version B.07.00, ESG_B Version B7.0, ESG_APDP Version B6.3) with the "Adjustments" radio button selected, and an E4436B instrument selected, the order of the tests is:
ANALOG BUS ADC
PRETUNE CAL.
VCO BIAS POT ADJ.
INTERNAL SOURCE CAL.
KV VS FREQUENCY CAL.
TIME BASE DAC CAL.
FM SCALE DAC OFFSET CAL.
.
.
.
etc.
Is this a problem with the software, or is the lock angle adjustment not necessary in the ESG-AP/DP instruments? I looked very carefully at the Calibration Guide and couldn't find anything that would indicate that this adjustment shouldn't or couldn't be done on the ESG-AP/DP instruments.
Is there a workaround? Is there a way to tell the software to run this adjustment on my E4436B? (For example, lying to the software by telling it that the instrument is an E4432B, or somehow installing the Lock Angle Pot test into the AP/DP "side" of the software.)
Regards,
Stan
这变得更加有趣!
锁定角度电位器调整肯定存在于软件的ESG-B端(例如,如果我告诉它我有E4432B),但ESG-AP / DP端不存在(当我告诉它时)
我实际拥有的内容:E4436B)。在我的服务支持软件(版本B.07.00,ESG_B版本B7.0,ESG_APDP版本B6.3)中选择了“调整”单选按钮,并选择了E4436B仪器,顺序为
测试是:ANALOG BUS ADCPRETUNE CAL.VCO BIAS POT ADJ.INTERNAL SOURCE CAL.KV VS FREQUENCY CAL.TIME BASE DAC CAL.FM SCALE DAC OFFSET CAL ....等这是软件的问题,或者是
ESG-AP / DP仪器中不需要锁定角度调整?
我仔细查看了校准指南,找不到任何表明不应该或不能在ESG-AP / DP仪器上进行此调整的内容。
有解决方法吗?
有没有办法告诉软件在我的E4436B上运行这个调整?
(例如,通过告诉软件该仪器是E4432B,或以某种方式将Lock Angle Pot测试安装到软件的AP / DP“侧”来说谎软件。)问候,Stan
以上来自于谷歌翻译
以下为原文
This is becoming even more intriguing! The Lock Angle Potentiometer adjustment is definitely present in the ESG-B side of the software (if, for example, I tell it I have an E4432B), but it is not present in the ESG-AP/DP side (when I tell it what I actually have: an E4436B).
In my Service Support Software (Version B.07.00, ESG_B Version B7.0, ESG_APDP Version B6.3) with the "Adjustments" radio button selected, and an E4436B instrument selected, the order of the tests is:
ANALOG BUS ADC
PRETUNE CAL.
VCO BIAS POT ADJ.
INTERNAL SOURCE CAL.
KV VS FREQUENCY CAL.
TIME BASE DAC CAL.
FM SCALE DAC OFFSET CAL.
.
.
.
etc.
Is this a problem with the software, or is the lock angle adjustment not necessary in the ESG-AP/DP instruments? I looked very carefully at the Calibration Guide and couldn't find anything that would indicate that this adjustment shouldn't or couldn't be done on the ESG-AP/DP instruments.
Is there a workaround? Is there a way to tell the software to run this adjustment on my E4436B? (For example, lying to the software by telling it that the instrument is an E4432B, or somehow installing the Lock Angle Pot test into the AP/DP "side" of the software.)
Regards,
Stan