大家好:我正在使用E5270B仪器对MOS器件进行基本的I-V测量(改变漏极电压和监测一个栅极电压下的漏极电流)。
与设定合规性不同,我想知道一旦测量的漏极电流值达到特定点,是否有任何方法可以停止测量,并且不会继续进一步继续直到达到用户提到的漏极电压的停止值。 谢谢,罗宾 以上来自于谷歌翻译 以下为原文 Hello All: I am doing basic I-V measurements (varying drain voltage and monitoring drain current at one gate voltage) on MOS devices using E5270B instrument. Unlike setting compliance I want to know if there is any way to stop the measurement once the measured drain current value reaches a particular point and not continue further till it reaches the stop value for drain voltage mentioned by the user. Thanks, Robin |
|
相关推荐
1个回答
|
|
您好Robin09,您可能会发现这个E5270B编程指南很有用:[http://cp.literature.agilent.com/litweb/pdf/E5260-90010.pdf] E5270B响应命令WM,其中,'WM'命令
将启用或禁用自动中止功能,该功能在合规事件发生时完全停止测试。 _例如:_ WM 2这将启用自动中止功能。 如果SMU在测试期间达到当前合规性限制,则立即停止测试。 WM 1如果SMU在测试期间达到当前的符合性限制,则电流将限制在极限,但测试将尝试继续,直到完成每个扫描参数。 IC-CAP并未“正式”支持此命令,但IC-CAP仪器的所有仪器选项都包含一个InitCommand选项,通过该选项可以向仪器发送额外的命令,驱动程序不会正式支持。 有时,这可用于实现驱动程序当前不支持的设置。 我目前无法访问实验室+设备进行全面测试,但一些初步测试表明在E5270B InitCommand字段中设置WM2,e..g输入:WM 2 n(发送命令并提供 新线)应该具有您正在寻找的效果。 IC-CAP仍然会发出有关所有值达到合规性的警告,但不应对每个点采取新数据。 请尝试一下,如果它有用,请告诉我们。 如果您有任何问题,请考虑联系Agilent EESof技术支持:[eesof_support@agilent.com]最好的问候 以上来自于谷歌翻译 以下为原文 Hello Robin09, You may find this E5270B programming guide useful: [http://cp.literature.agilent.com/litweb/pdf/E5260-90010.pdf] The E5270B responds to a command WM, where, the ‘WM’ command will enable or disable the automatic abort function, which completely stops the test in the event of a compliance event. _For example:_ WM 2 This enables the Automatic Abort function. If the SMU hits the current compliance limit during the test, the test is immediately stopped. WM 1 If the SMU hits the current compliance limit during the test, the current will clamp at the limit but the test will attempt to continue until it finishes each of the sweep parameters. IC-CAP does not 'formally' support this command, but all Instrument Options for IC-CAP instruments include an InitCommand option via which one can send extra commands to the instrument, that the driver does not formally support. Sometimes this can be used to achieve settings which the driver does not currently support. I do not currently have access to a lab + device to a full test on this, but some preliminary tests suggest that setting WM2 in the E5270B InitCommand field, e..g enter: WM 2n (to send the command and provide a new line) should have the effect you are looking for. IC-CAP will still send warnings about having hit compliance for all the values, but new data should not be taken for each point. Please try that and let us know if it is helpful. Please also consider contacting Agilent EESof Techical Support for questions like this: [eesof_support@agilent.com] Best regards Else 附件
|
|
|
|
只有小组成员才能发言,加入小组>>
1298 浏览 0 评论
2378 浏览 1 评论
2199 浏览 1 评论
2068 浏览 5 评论
2955 浏览 3 评论
1142浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
760浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
951浏览 0评论
1299浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-29 06:40 , Processed in 1.245159 second(s), Total 48, Slave 42 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (威廉希尔官方网站 图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号